Plugged columns!!!

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi! If someone could answer why I have this problem with different kinds of columns when mobile phases contain sodium salts of butanesulfonic or heptanesulfonic acid. I have used a Bondclone 10µ C18 with sodium butanesulfonate 0.006M with 5% methanol and the pressure kept increasing as weeks went by until I could not use it anymore and the same is happening with a second column.
Something similar happened when I used a Spherisorb ODS-2 3µ with sodium heptanesulfonate 0.03M with 5% methanol.
In both cases I used a flow of 2.0 mL/min and I also cleaned the columns after each and every run, first with water and then with a high proportion of organic solvent (acetonitrile in most cases).
I have tried washing them with 100% organic solvent at a low flow but the problem remains... Please help me...!

Dominique.

I do not think that the salts are to blame (although 30 mM of heptasulfouric acid starts to be pretty high, are you sure is not 3 mM?). Probably the problem comes from what you inject (i.e. your sample; what is it by the way?).

Hi!
I am analyzing medicines which do contain any number of excipients. I did consider that maybe some matrix component is retained in the column, but I have washed the column with tetrahydrofuran and it has made no difference, so I don't know what to think anymore. I'm afraid to damage the column beyond repair if I go more hydrophobic with the solvents... And yes, I also thought that the salt concentration of 30 mM was terribly high plus i forgot to mention that this mobile phase also has orthophosphoric acid and triethylamine. I don't have much liberty for method development and modifying, but i do try to take care of my poor columns as much as I can.

I think I would try flushing the columns initially with a water/acetonitrile mix that you know all of your sample components are soluble in, rather than pure water. I would also still go to the high acetonitrile flush, but perhaps even initially with a little acid to help move junk along.

I assume you are using HPLC grade reagents, if not - you should consider filtering the mobile phase, and perhaps filtering or centrifuging your samples. I would also invest in a guard column, if you're not already using one.

Please keep having fun,

Bruce Hamilton
4 posts Page 1 of 1

Who is online

In total there are 24 users online :: 2 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Google [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry