Unusual repeating peaks in blank
Posted: Wed Aug 28, 2019 2:33 pm
				
				I am having an issue with my GC-MS where I have repeating peaks in my baseline. The attached images are from no-injection blank runs with a gradient from 100 to 200 C; however, this occurs under isothermal conditions as well. 
The second image shows the difference at different flow rates where the time delta between peaks is very similar (~0.14 min).
The mass spectra of these peaks are showing a primary peak at 32 with a second peak at 40 (both show spikes in their respective EICs).
I've tightened all connections and "baked out" MS. The air/water check in the MS is all fine (H2O: 1.2%; O2: 1.0%; N2: 3.3%)
Has anyone encountered this before?

Different flow rates (0.8 & 1.2 mL/min)

			The second image shows the difference at different flow rates where the time delta between peaks is very similar (~0.14 min).
The mass spectra of these peaks are showing a primary peak at 32 with a second peak at 40 (both show spikes in their respective EICs).
I've tightened all connections and "baked out" MS. The air/water check in the MS is all fine (H2O: 1.2%; O2: 1.0%; N2: 3.3%)
Has anyone encountered this before?
Different flow rates (0.8 & 1.2 mL/min)