Sadly, Uracil is very, very poorly retained, and is used by most manufacturers as the void volume marker because it's easily detected. It would not be a good compound to test the column's condition, and you have confirmed your instrument's flow rate is OK ( which void volume might confirm ).
I would suggest that if you can't find any of the retained components of the manufacturer's test mixture, then try find any sample that was analysed using a similar method before the first sample was analysed, and repeat that analysis. Also repeat the first sample analysis just to ensure that sample also now elutes early.
One mistake I routinely make following published methods is that I prepare 60:40 CH3CN:H2O instead of H2O:CH3CN, however the pressures would be significantly different.
I suspect that your column may have died if the pressure are similar to the first run, because pressure is usually a good indication of wrong solvent or precolumn leaks. If you had a leak between the column and detector the first time, you would get longer retention without a noticeable change in pressure.
I'd definitely really want to try and obtain at least one retained component of the manufacturer's test mix, because that remains the fastest way to confirm the column has changed/died since manufacture.
However if you are desperate, there are some published retention volumes for acetone, heptanone and benzene on a Curosil PFP column in a published thesis ( Fig 3.37 p152 onwards ) if your eyes don't glaze over finding that data at the location below. Link may not work, then search using Google.
http://domapp01.shu.edu/depts/uc/apps/l ... 5.pdf?Open
Please keep having fun,
Bruce Hamilton