I've developed some HPLC methods for an OTC drug which contains Acetaminophen and Ibuprofen. I've used gradient elution in its determination since those two components have different polarity which is quite difficult when using an isocratic. The condition is as shown below:
Column: C8 encapped
Mobile Phase: ACN (A) : Monobasic Potassium Phosphate 0.025 M, adjusted with phosporic acid ad pH 3.0 (B)
Gradient program : 0 min: 2% A (hold for 9 min); 10 min: 50% A (hold for 10 min); 21 min: 2% A (hold for 7 min --- reequilibrium)
Flow rate : 1.5 ml / minutes
Wavelength : 250 nm and 223 nm
I've got such a good separation with the condition above but there was a problem with the baseline. When I run baseline, mostly before the analysis was started, there was a ghost peak at the retention time of Ibuprofen. I've changed the composition of mobile phase while it was running at high organic content (after 10 minutes run) as well as filtering the buffer through column before its using but all of them seemed not quite useful. The ghost peak remained to be appear at the same retention time of Ibuprofen.
I need some suggestions and considerations to obtain a better determination method.