TMC » Tue Dec 23, 2008 8:45 pm
I am having a new problem with an ELSD lipid method that I have been developing. The problem is that recently matrix blanks and samples are producing ghost peaks while diluent blanks are not. The retention times and areas of the ghost peaks are not reproducible from one injection to another. Here is some background information. The diluent is a mixture of hexanes/IPA/water. The matrix blank is 10 % of 400 mM sucrose/20mM phosphate pH 7 in water and 90 % diluent. Samples are centrufuged at 16000 rcf for 30 minutes. HPLC method is 0.7 mL/min 0.1 % TFA in 80 % organic (isocratic at start followed by gradient). Ghost peaks are occurring in both isocratic and gradient parts of the method and have a height of about 1-2 LSU units. ELSD conditions are 25 psi nitrogen, 35 C nebulizer, 50 C drift tube. Injection volume is 30 uL. I have tried to isolate the problem by changing each one of the following variables one at a time:
Cleaned nebulizer with methanol then water, cleaned the drift tube with heated water, removed guard column, tried an older column, new mobile phases, extend the high organic washout, extend the re-equilbration time at the end of the method.
This is a relatively new column (<100 injections) and the method I am using has not changed from when I had no ghost peaks.
Any ideas out there?