First off, I know I have an injection volume issue, as I'm using a GC PAL autosampler, and the syringes we've been getting are extremely substandard. It goes primarily to the manufacturer of the syringes, and the area reproducibility for a given peak/standard has a %RSD of 10% on ten replicate injections of the same standard. Obviously not good at all. But with that said, I am using on-column injection with very simple non-polar compounds on a simple non-polar column, and I would expect area ratios to not be a problem. They are.
Injection solution is ~ 1000 ug/L each in C11 and C22 alkanes, dissolved in MeCl2. Injection volume is 1 uL.
Instrument is HP5890 Series II, with EPC, and cool on-column injection.
Column is 20 m x 0.53 mm DB-5, with 1 um film thickness.
Carrier is He @ 10 cc/min
Inlet is programmed as
60 C initial temp., held for 0.5 min.
Ramped to 320 C @ 100 C per min., with zero hold time
Oven is programmed as
60 C initital temp., held for 1.5 min.
Ramped to 320 C @ 10 C per min., no hold time
Detector is FID @ 320 C
Detector gases are He makeup @ 20 cc/min., H2 @ 30 cc/min., Air @ 400 cc/min.
(I should note that I've been using this instrument for years, with similar ramps and rates, flows, etc. without problems. I like to use the on-column inlet with the thin film non-polar megabore columns because I often analyze TMS derivatives of wide molecular weight range samples vs. internal standards.)
My peak shapes are excellent. Visually everything looks very good. However, the area ratios of the C22 peak to the C11 peak are as follows;
0.890546373
0.849581975
0.857527365
0.843012247
0.824673196
0.833617403
0.917500427
0.923461612
0.862258644
0.91404731
(please forgive the inordinate amount of sig figures)
The %RSD of these is > 4%. As I mentioned earlier, I know I have non-reproducible injection volumes. However, the error I see there should not come into play for area ratios. Also, flash vaporization should not be an issue with this system in the cool on-column mode. That's the beauty of the on-column design, as long as the initial inlet T is appropriate.
The construction of the inlet for a 5890 on-column system really doesn't lend itself for compounds to be "lost".
There is no undue noise in the baseline. Retention times are reproducible.
Does this seem like a FID problem? Or perhaps an electrometer or main board problem? The instrument is quite old (~ 15 years) but the beauty of some of these 5890's is that they seem to run forever.
I don't think I've overlooked anything obvious, but if I have I'm more than willing to accept that I'm wrong!
Any thoughts or replies are greatly appreciated.
Best Regards to all.