						I'm running an isothermal method looking at crude oil from C6-C20 on a ZB1-inferno (30x0.25x0.25), 150 split ratio.  Identifying the n-alkanes is easy.  They're obviously visible.  But consistent ID of inter-alkane peaks becomes difficult, particularly after system maintenance.  I've tried calculating retention indeces.  I say plural because I've done Kovats, Linear, Linear interpolation, cubic spline, bessel spline, etc.  When I calculate them for the same peaks on a run of the same sample after inlet maintenance (column clip included), the indeces for most peaks is fine, (+/- 0-2 points).  But for the heavier inter-alkanes, the indeces shift upwards of 8 points.  Is this a "normal" phenomena?  Does this mean there is something "wrong" with my system?  Do I need to change a setting in my method?  Being able to consistently id these smaller peaks from one sample set to the next is critical to my experiments and any assistance would be GREATLY appreciated!
					
									- 
								
 - Posts: 3477
 - Joined: Mon Jan 07, 2013 8:54 pm
 
When you clip the column are you adjusting the head pressure lower to account for the shorter length so you have the same linear velocity through the column?I'm running an isothermal method looking at crude oil from C6-C20 on a ZB1-inferno (30x0.25x0.25), 150 split ratio. Identifying the n-alkanes is easy. They're obviously visible. But consistent ID of inter-alkane peaks becomes difficult, particularly after system maintenance. I've tried calculating retention indeces. I say plural because I've done Kovats, Linear, Linear interpolation, cubic spline, bessel spline, etc. When I calculate them for the same peaks on a run of the same sample after inlet maintenance (column clip included), the indeces for most peaks is fine, (+/- 0-2 points). But for the heavier inter-alkanes, the indeces shift upwards of 8 points. Is this a "normal" phenomena? Does this mean there is something "wrong" with my system? Do I need to change a setting in my method? Being able to consistently id these smaller peaks from one sample set to the next is critical to my experiments and any assistance would be GREATLY appreciated!
The past is there to guide us into the future, not to dwell in.
				- 
								
 - Posts: 1773
 - Joined: Fri Feb 09, 2007 10:33 pm
 
						I am curious, and it really is none of my business, but why are you doing isothermal between C6 and C20??
Best regards,
AICMM
									Best regards,
AICMM
- 
								
 - Posts: 7
 - Joined: Mon May 15, 2017 2:25 pm
 
To clarify, my temperature ramp is linear. I don't know why I said my method was isothermal. Sorry.I am curious, and it really is none of my business, but why are you doing isothermal between C6 and C20??
Best regards,
AICMM
- 
								
 - Posts: 7
 - Joined: Mon May 15, 2017 2:25 pm
 
No... ?When you clip the column are you adjusting the head pressure lower to account for the shorter length so you have the same linear velocity through the column?I'm running an isothermal method looking at crude oil from C6-C20 on a ZB1-inferno (30x0.25x0.25), 150 split ratio. Identifying the n-alkanes is easy. They're obviously visible. But consistent ID of inter-alkane peaks becomes difficult, particularly after system maintenance. I've tried calculating retention indeces. I say plural because I've done Kovats, Linear, Linear interpolation, cubic spline, bessel spline, etc. When I calculate them for the same peaks on a run of the same sample after inlet maintenance (column clip included), the indeces for most peaks is fine, (+/- 0-2 points). But for the heavier inter-alkanes, the indeces shift upwards of 8 points. Is this a "normal" phenomena? Does this mean there is something "wrong" with my system? Do I need to change a setting in my method? Being able to consistently id these smaller peaks from one sample set to the next is critical to my experiments and any assistance would be GREATLY appreciated!
- 
								
 - Posts: 3477
 - Joined: Mon Jan 07, 2013 8:54 pm
 
Most systems with electronic flow control have to calculate the flow using the column internal diameter and length to know what head pressure will deliver a certain flow rate (ml/min or cm/sec whichever it is set to use). The shorter the column the less head pressure required to achieve the desired flow. If you trim column but do not tell the instrument the new length, or manually adjust the head pressure if not using electronic flow control, then the controller will apply too much head pressure for the flow required. Also I don't believe it is a linear relationship as temperature increases, so you may have 1.2ml/minute initial flow when you expect 1.0ml/minute and maybe 1.3 or 1.4ml/minute at maximum temperature when you expect 1.0ml/min. If that is happening it could be throwing off the later eluting analytes more than the early analytes.No... ?When you clip the column are you adjusting the head pressure lower to account for the shorter length so you have the same linear velocity through the column?I'm running an isothermal method looking at crude oil from C6-C20 on a ZB1-inferno (30x0.25x0.25), 150 split ratio. Identifying the n-alkanes is easy. They're obviously visible. But consistent ID of inter-alkane peaks becomes difficult, particularly after system maintenance. I've tried calculating retention indeces. I say plural because I've done Kovats, Linear, Linear interpolation, cubic spline, bessel spline, etc. When I calculate them for the same peaks on a run of the same sample after inlet maintenance (column clip included), the indeces for most peaks is fine, (+/- 0-2 points). But for the heavier inter-alkanes, the indeces shift upwards of 8 points. Is this a "normal" phenomena? Does this mean there is something "wrong" with my system? Do I need to change a setting in my method? Being able to consistently id these smaller peaks from one sample set to the next is critical to my experiments and any assistance would be GREATLY appreciated!
The past is there to guide us into the future, not to dwell in.
				- 
								
 - Posts: 62
 - Joined: Sun Feb 17, 2019 11:50 pm
 
What's your GC detector?I'm running an isothermal method looking at crude oil from C6-C20 on a ZB1-inferno (30x0.25x0.25), 150 split ratio. Identifying the n-alkanes is easy. They're obviously visible. But consistent ID of inter-alkane peaks becomes difficult, particularly after system maintenance. I've tried calculating retention indeces. I say plural because I've done Kovats, Linear, Linear interpolation, cubic spline, bessel spline, etc. When I calculate them for the same peaks on a run of the same sample after inlet maintenance (column clip included), the indeces for most peaks is fine, (+/- 0-2 points). But for the heavier inter-alkanes, the indeces shift upwards of 8 points. Is this a "normal" phenomena? Does this mean there is something "wrong" with my system? Do I need to change a setting in my method? Being able to consistently id these smaller peaks from one sample set to the next is critical to my experiments and any assistance would be GREATLY appreciated!
			7 posts
							Page 1 of 1
					
	Who is online
				In total there are 23 users online :: 2 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: Google [Bot], Semrush [Bot] and 21 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Semrush [Bot] and 21 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
 - Follow us on Linkedin: Separation Science
 