- Posts: 129
- Joined: Tue Dec 04, 2007 12:55 pm
The USP <467> specifies a flow rate of 35 cm/sec for the determination of residual solvents.
Does this mean the linear velocity is always 35 cm/sec during the temperature ramp?
Terry
Advertisement
Discussions about GC and other "gas phase" separation techniques.
I am interpreting (and working) this like: use constant linear velocity of the carrier gas at 35 cm/s throughout the chromatographic run (working with constant linear velocity of the carrier). When you set your PPC to work with constant linear velocity it compensates the variation of the viscosity of the carrier during the temperature ramp, increasing the column head pressure (when T raises) in order to maintain constant linear velocity of the carrier gas in the column.Hi, there
The USP <467> specifies a flow rate of 35 cm/sec for the determination of residual solvents.
Does this mean the linear velocity is always 35 cm/sec during the temperature ramp?
Terry
According to ideal gas equation PV=nRT, the gas volume is proportional to the temperature while keeping the pressure constant. So, if theThanks for all your prompt reply.
As Krickos said, I am running an Agilent system, 6890N.
At 40°C, 35 cm/sec helium through a 30m, 0.32mm ID column has a flow rate of 2.2 ml/min; at 240 °C, helium flow rate would be 1.5 ml/min with the same linear velocity.
As far as I am concerned, carrier gas linear velocity is proportional to flow rate, and is independent of gas viscosity (oven temperature).
Constant linear velocity means constant flow, right?
Terry
Terry, this "As far as I am concerned," is a new scientific method?
Peter, was a bit unclear there and more or less used a "backbone" recommendation. Broke could be either column and phase but hardly a risk in this case as discussed above. Seems to remeber that manufactures of columns like HP used to have some recommended pressure intervals for capillary columns. We run almost exclusively capillary columns here but also a few PLOT columns so "watch the pressure" tends to be includedKrickos, is it really possible to burst a silica column by the gas pressure in a GC ? I wonder what the burst pressure really is - anyone want to speculate ?
Peter
Unfortunately, there is no 'Constant Velocity' mode in the Chemstation. So I would have to chose either 'Constant Pressure' or 'Constant Flow'On the instrument that you have you can set the linear velocity, and the software calculates the volume flow, and adjusts it during the temperature programme to keep the linear velocity constant. You need to use this feature and not worry about setting and programming the volume flow rate.
I have used this feature in B.02.01 Chemstation, it is fun. But in B.03.01, it is not available any more.Some detailed info. If you use chemstation you swith the setting under columns in the "mode" window. Per default in chemstation you have above column, oven etc a plot of GC oven temperature versus time. To the left of the plot area you have a PLOT button, pushing enables you to add more info to the plot, you can simultanious plot column, flow/velocity/pressure to verify your settting or just for fun/info