Trimethylamine or TEA
Posted: Mon Aug 12, 2019 12:20 pm
by ydna1977
Hi
Which is the suitable Ion pair reagent to use ?
What difference would each Ion pair usually make on peak separation ?
Ion pair will be combined with buffer and SLS for mobile phase with methanol
thanks
Re: Trimethylamine or TEA
Posted: Mon Aug 12, 2019 3:15 pm
by DR
Ion pairing agents are usually chosen at least in part as a function of the pH at which you need to run. Ideally, you choose a pairing agent that will afford you at least a little buffering capacity at your desired pH.
TEA is a common abbreviation for Triethylamine.
Did you mean to ask about whether to use TEA or TFA (Trifluoroacetic acid)?
If so, TFA is used at much lower pH than TEA.
Posted: Tue Aug 13, 2019 4:35 am
by Consumer Products Guy
In the consumer products field TEA is a nickname for triethanolamine.
Posted: Tue Aug 13, 2019 10:58 am
by ydna1977
No
asking about what is ' best' to use Trimethylamine or Triethylamine
Posted: Tue Aug 13, 2019 2:49 pm
"Best" is what works best for your individual situation. You determine that, that's what they pay you for.
Posted: Tue Aug 13, 2019 7:35 pm
by HPLCaddict
Trimethylamine: gaseous
Triethylamine: liquid
I'd prefer to work with the liquid additive:)
Posted: Tue Aug 13, 2019 7:39 pm
by HPLCaddict
"Ion pair will be combined with buffer and SLS for mobile phase with methanol"
SLS : sodium lauryl sulfate? That's your ion pairing additive!
Both TEA and TMA are no real ion pairing additives. Why do you want to add one of them at all? TEA was used in the old days for masking unwanted silanol interactions and thus improving peak shape. With modern, high purity silica columns, it's seldom needed anymore.
Posted: Wed Aug 14, 2019 8:02 am
by ydna1977
yes SLS ion pair and guess TEA or TMA used to decrease interactions and suppress tailing .. though on C18 column with this mobile phase I am not sure
Posted: Thu Aug 15, 2019 4:04 pm
by DR
Clearly, I missed an "m".
Apologies.
Posted: Thu Aug 29, 2019 7:28 am
by ydna1977
Found TMA 33% in ethanol from a supplier as a replacement for just TMA
selectivity much improved compared to using same volume of TEA, so not just for improving peak shape here ?