I have just started on a PhD in chemical engineering, and my group are working on catalytic conversion of biomass to biofuels. One of our catalysts converts cellulose to ethylene glycol/polyethylene glycol and other -diols, both straight and ring formations. We also have some HMF derivates such as tetrahydrofurfuryl alcohol.
Setup:
Agilent 1260 HPLC outfitted with a HiPlex Ca-duo column
Agilent 6120 Quadrupole LCMS System (ESI ionization)
As of now, we have just used distilled water as our mobile phase. No buffer or anything added to it.
Questions: Is it possibe to remove the +Na adduct all together? We tried using 0.1% Formic acid, but to no avail. The baseline noise just increased and some products where not found using this mobile phase. The ONLY ion we see in our standard solutions is the [M+Na]+. No real fragments occuring. The peaks are also pretty broad.
However, when analyzing our products, we get several peaks that can be described by different fragmentations such as M+Na-H2O+H, M+Na-OH+H etc..
Is this ok to use for characterization alongside results from GC-MS? The problem is that we have peaks in the GC-MS that are not easily characterized, and we have the same problem in the HPLC.
So, should we try to remove the Na+ or just run with the ions and fragments we get at this point? Will changing dest.water to HPLC grade water or installing a desalter column between mobile phase and HPLC or between the HPLC and MS help at all?