I'm currently putting a method together for derivitised phenols from water. It's a relatively large suite of 27 phenols from phenol up to Dinoseb.
The problem I'm having is derivitising the phenols after extraction. I believe I have enough checks in place to be confident that the extraction and analytical phases are working.
Here is a brief outline of the extraction.
100ml of sample is acidified with HCl to pH 2 then passed through a conditioned SPE cartridge.
The cartridge is left to 'dry' with full vacuum on for 15-20 minutes.
The column is then eluted with acetone followed by ACN.
The extract is passed through a column containing magnesium sulphate (anhydrous) before derivitisation with BSTFA/TMS
The final concentration after extraction is 1000ppb of each phenol.
1ml of the final extract is then spiked with 50ul of Int Std (including bromophenol to check effectiveness of derivitisation) followed by 100ul of BSTFA/TMS.
The reason I am confident that the extraction is working is that if I do a X10 dilution on the extract the derivitisation works perfectly, but on the undiluted extract derivisation doesn't work at all, as confirmed by the addition of the bromophenol after extraction.
I believe the problem is residual water inhibiting derivitisation but none of the techniques I have tried so far have helped.
So far I have tried:-
Leaving the column with vacuum for 1 hour
Drying the column in 60 Deg C oven for 30 mins before elution
drying extract in sodium sulphate and magnesium sulphate.
Any bright ideas out there ?