Make up gas showing a flow when there is none.

Posted: Thu Aug 08, 2019 10:09 am
by Chemiger
hello,
I am currently using a 6890
the make up gas flow is showing a value of 10 when the GC is running even though there is no make up gas flowing.
Is this normal ?

Thanks,
Chemiger

Re: Make up gas showing a flow when there is none.

Posted: Mon Aug 12, 2019 2:45 pm
by tkubowicz
Hello

You didn't specify detector but I assume you mean FID.
Just disconnect makeup gas line from GC (at the back) and check flow on GC display. Check value 0, check value 10, 25,50 ml/min. If you still have value 10 displayed it means that EPC module could be broken.
Of course make sure that you disconnected correct line (if you have 2 detectors).
I'd strongly recommend to check actual flows with flowmeter (you need to turn off H2 and Air).

Regards

Tomasz Kubowicz

Re: Make up gas showing a flow when there is none.

Posted: Tue Aug 13, 2019 7:58 am
by dblux_
Not a bad idea to zero pressure sensors.