How do I add a line in masslynx and keep on running
Posted: Tue Aug 06, 2019 9:19 pm
by wout
How can I simply add or insert a line in a sample list in de masslynxsoftware during a run? And being sure that the inserted line is going to be analyzed?
So it's about editing a running sample list.
Should I save (again) the customized sample list?
Should I start again the customized sample list or do nothing?
Or should use the PAUSE button, insert a line and then restart the sample list?
As you can see I'm not sure what is the best way to do this
Re: How do I add a line in masslynx and keep on running
Posted: Wed Aug 07, 2019 9:46 am
by Aldijkhuizen
Pause the running sequnce
Alter the seqience
Save sequence
Start swquence from next injection
After new seqience start delete the paused sewuence
Posted: Tue Aug 13, 2019 9:10 am
by lmh
With MassLynx to be honest I'd personally recommend you never modify any part of a sample list that you've submitted to be run. By all means add new lines to the end, but editing the list in the middle can often cause terrible confusion.
In contrast to most manufacturers, the Waters sample list is what it says it is: just a list. The editor is just an editor. It might look like Agilent's sequence table or Shimadzu's batch file but it isn't. You put samples on the list, and you tell MassLynx to run rows from the list. Viewed that way, it's quite a nice concept.
For this reason, you don't need to pause the instrument when you edit the list. You can edit it at any time, adding new lines, but even if you save the list afterwards, they won't be run. They will only be run if you tell MassLynx to run them, for example by selecting them and clicking the "run" button as normal. The rows that you have submitted will now appear in the queue.
If you realise you've submitted inappropriate samples to be run, what I'd recommend is write everything the way you want it in the sample list, submit what you've written, and then delete the current item from the queue. This way, MassLynx can finish the current run and simply proceed to your new lines without confusion. The place where things get horrible is if you delete a line from the middle of a set of samples that's already been submitted; I suspect MassLynx of getting itself into a sort of frame-shift misunderstanding here where it injects samples off one line while using the name off the adjacent line, but it's probably just that I misunderstand the logic behind how it works. I asked a Waters engineer about this and his response was "just don't do it; it's too confusing!"
Posted: Wed Aug 14, 2019 3:04 pm
by wout
Right Imh,
I think you are right. I tried the pause button but then you have to be sure that the existing line is completly finished and the pause moment is started.
That's complicated. Better do it your way and maybe start another shorter sample list with alle the samples you want to add. That always works.
But thank you for answering my question.