- Posts: 480
- Joined: Fri Sep 03, 2004 7:57 am
I need suggestion for my developed method to separate an active pharmaceutical ingredient with its main degradation product.
The formula of the api:Ar-CONH2, and the degradant: Ar-COOH. Both are very polar compounds.
The latest chromatographic condition:
Column : Atlantis dC18 (5 μm) 150 mm x 4 mm
Mobile phase: 25 mM (NH4)2PO4 , 0.1% TEA, adjusted to pH 7.0 with 10% H3PO4 .
Oven temperature: 25°C
Flow rate : 0.5 ml/minute
On that condition, I get results:
Rt Ar-COOH = 12.807 minutes
Ar-CONH2:
Rt = 13.893 minutes
Resolution = 1.67
k’ = 6.82
Asym (10%)= 1.38
Tp = 7096
Decreasing pH of mobile phase up to 2.5 will decrease the resolution.
How to get more adequate condition for this method? My targets are better resolution and shorter run time.

Best regards,
SYX