My sample readings on different pH instrument are showing variation of around 0.3 whereas my standards in the same range showing readings within ±0.03
I have tried on different makes of pH meter and electrodes although I am getting this variation in readings of only samples not for standards.
Why this variation is there for only samples( Injectable product)
Sitaram B.

Did you calibrate the different pH meters and different electrodes with your standard(s)?

I am not an expert in pH measurements but are the ionic strenths of your samples much lower than the standards?
Very low ionic strengths can cause rather large variations in pH readings. If this is the case addition of a neutral electrolyte (like sodium nitrate or sodium bromide) might help.
--
Robert Haefele
Thanks for reply
Sitaram B.
4 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry