Corona Detector Troubleshooting

Posted: Sat Aug 03, 2019 3:06 pm
by Vlad Orlovsky
Hello Everybody,

I need help figuring out how to troubleshoot Corona problem. I am trying to set up this detector and it fails self calibration and gives error 80. You can skip it and attempt to run it gives error "low ion trap voltage". As a remedy, it suggests to contact manufacturer. May be somebody knows how to address this issue. It is used Corona CAD.

Thank you very much

Re: Corona Detector Troubleshooting

Posted: Thu Aug 08, 2019 11:39 am
by Perreman
Perhaps you should try contacting Thermo about this issue. Of course they will probably not help you for free, but it sounds like an issue where you might need some part(s) replaced in the CAD.

Re: Corona Detector Troubleshooting

Posted: Tue Nov 03, 2020 6:49 pm
by LCwatcherGE
Dera Vlad,

the CAD ist flooded. You'll need a SVC Rep onsite to perform drying, cleaning and PM.

Re: Corona Detector Troubleshooting

Posted: Thu Nov 12, 2020 12:48 pm
by BostonFSE
Error code 80 means its flooded. Liquid flow was sent to the CAD with the gas OFF. The gas should always stay on when liquid is flowing into the CAD. Some people use the UV only and forget about the CAD and when the gas runs out, it can flood. At a minimum, it needs a PM.