Internal standard increases, analytes decrease in area......
Posted: Tue Dec 23, 2008 11:19 pm
I've read the forums for a while and decided that I'd post the problem in the hopes that someone has had a similar experience.
I'm using a micromass quattro LCZ older instrument I know but it's what I have. 2795 waters HPLC on the front end and I've done an injector test and there is little to no injector problem. Running MRM experiments.
When the DMPK group runs samples, ~300 a night, mixture of qc, analyte with internal standard, washes...etc., the internal standard starts out at an area count of, say, 4000, and ends up at a area count of 7000 or so, while in the same samples the analytes start at around 6000 and drop to about 4000 or so. QC samples Bear this out and have same response trends.
Anyone ever seen this before? IS going up while Analytes going down over time?
Just had the instrument PM'd but it was doing it before. The sensitivity is up now and that's good. Using a diverter valve so no goop getting in my source(at least not like before). the run in 5 minutes long and everything comes out "chromatographically" nice, no peak overlaps. (~20s separation or so)
Driving me crazy and any help would be appreciated.