A very broad hump in gradient HPLC can result from two things:
- build-up on the column of contamination in the weak solvent
- RI mismatch between the solvents.
As XL suggested, if the size of the hump varies with equilibration time, then it is contamination. If it does not vary with equilibration time, then it is probably a refractive index effect (RI is a non-linear function of composition, which is why the baseline shift can appear as a "hump". If the hump is very broad, your best approach is to simply ignore it and let the data system deal with it (in the vicinity of the peak, the data system should treat it the same way as it would treat drift).
All UV detectors show some response to RI, but iff the hump is excessively large, try to run the same method on a different instrument of the same model (if available). If both systems show about the same amount of hump, then it's an optics design issue, and there's not a lot you can do about it chromatographically. If one of the systems looks worse than the other, then it is a service issue; if the flow cell is poorly aligned with the optical axis of the detector, proportionally more light enters the cell "off axis", and RI sensitivity is exacerbated.