I have been running GC/MS for over 3 years but am new to running splitless. Currently I am using straight liners packed with glass wool in the middle. I am having troubles with gigantic peaks coming out after the solvent front should be alluting and drowning out some of my analytes. I am also finding that when I am changing liners, the glass wool ends up right at the top of the liner (right below the septum, basically), rather than in the middle where they start out (when I install them). I am still trying to figure all this out. Does this mean anything to anyone else?
Thanks.
Well could be some other issues with regard to the giant peak, but you have not posted any details on what you run and how the instrument is set up.
As for you choice of liner: Not good, for splitless I use a tapered liner (narrow at bottom or double tapered narrow at bottom/top), secondly I tend to use "focused" liners ie they are made so that the prepacked wool stays in place especially for splitless injections.
There reason for the wool moving around could be that you pack too little if it alternatively too much so the syringe gets "stuck" in it and pulls it upwards.
The wool issue is of course one possible reason for the "giant" peak issue.
Also I would typically turn on the split perhaps 2min or so depending on settings after each injection to "clear" the injector.
Using a straight liner in splitless may led to that the evaporated sample escapes from the liner and enters the injection port area and in worst case the gaslines.
Please post more details so we better can help you.