krickos » Fri Dec 05, 2008 1:16 pm
Hi
Would not say it is just the FID. More the headspace technique itself and the method in use in combination with the low levels your looking at.
A total of 0,1ug in the vial is hard to analyse but a rough approximation could be: 50% of 0,1ug acrylonitrile ends up in 10ml gas phase, 1ml is injected with a split ratio of 1:5=> 1 ng (1000pg) on the column. As a comparision you typically can detect about 20pg toluen (that ends up on capillary column) with a FID. Sure there is a difference in the responsfaktor between the two compunds but roughly.
So purely teorethical you should be able to detect it if you can maximise the acrylonitrile concentration in the gasphase and minimise the split ratio.
A few possible ideas to try on a larger concentration to see any enhancement:
1. You refer to an application, did it work and have you done something different?
2. Maximising gasphase concentration.
You are not solving the sample in water so try raising HS oven to 100°C (over bp of acrylonitrile). In water solutions you may use "salting out" technique ie adding a salt like sodiumsulfate to push a polar solvent into gasphase but will likely not help as much here.
Is it possible to increase sample amount?(ie increasing spike)
3. Minimise split ratio. Using an Agilent instrument you typically need to stay on a total flow of 10ml/min (carrier+split).
4. Minimise risk of active surfaces. acrylonitrile is polar so if you use a valve/loop HS make sure that loop/valve/Transfelline is at least 100°C to avoid loss/interactions of analyte.
5. Vial pressurazation. Again mayby depending on what system you use but if you use an Agilent 7694 or the new G18888 you may want to cosider to leave the pressure on but setting time to Zero. At high temperatures the vial pressure should be enough to flood the lines/loop. Pressarusation or to long time may cause a loss/dilution of analyte (leaks etc).
Unsure if there are any specifics you need to pay attention with regarding acrylonitrile have not worked with it, but actetonitrile I have managed to detect at low ug's with a "standard method" (a total of 9ug in 4ml DMF in a 10ml vial, HS oven 70°C, split 1:7).
Seems like you will need to work hard to obtain that sensitivity.