We worked on separation of both organic and inorganic anions on Acclaim Mixed-Mode columns using ELSD. Currently, we developed a new stationary phase which can separate both anions and cations within the same analysis. Some aspects were presented in the following poster:
http://www.dionex.com/en-us/webdocs/675 ... l-note.pdf
(Please pay attention to Figures 6 and 7).
Analysis of inorganic anions by ELSD shouldn't be a problem when the analytes are above 10 ppm. Depending on samples, mobile phase conditions, ELSD setting and the make of ELSD, the sensitivity may vary significantly. For NaCl standard in rather optimized conditions, the detection limits of Na+ and Cl- I could get were 1 ng and 10 ng, respectively.
ELSD is a universal detector (not selective). For high sensivity and high selectivity analysis of inorganc anions and cations, ion-chromatograpgy is a way to go.
Regards,