Problem with the MS 5973: Difficulty with the EM supply

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hi people!

I work at a GC/MSD (6890/5973).
This system don’t has an UPS, and after some interruptions of electrical supply that began to fail.
This is the error message in the Logbook:

1101, "[{REP} acq: strt sim]: Unrecoverable mass spec fault: 512"
Difficulty with the electromultiplier supply at runtime 7.44

This problem had happened beforehand, but the failure was in the method, since power failures ocurred during a run, and the method was corrupted.

Today, I executed a run, and the next run stopped at 7.44 minutes.

The filament is good and the vacuum is good too. The EMV was 1500.

I executed Profile and Scan (in Manual Tune) and the error message is the same.

Help me please!

Astroguille,

Probably is worth to use a different method to see if you get
the same error message and also to retune the instrument using a
previously saved copy of the autotune macro (or a macro from another
instrument or computer).

As you mentioned something similar already happened in the past...



Best regards,
bhuvfe
2 posts Page 1 of 1

Who is online

In total there are 42 users online :: 1 registered, 0 hidden and 41 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Majestic-12 [Bot] and 41 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry