Hi guys,
I think that I could have used a better phrase than 'gunk'. Although my work up until now has focused on known analytes (primarily organic acids), I'm getting a bit more free-form in the holiday season and I'm stepping back and using my QQQ for a few metabolomic experiments. With my really clean samples (etoac xn), the baseline is often minimal, even in increasing organic... until I increase the flow rate to shorten analysis time. The peaks that I get as the baseline climbs are still of interest, but I wonder if I should consider using different scan rates as things do co-elute. Perhaps this is an empirical question and I should just do the experiment (with the help of something like mzmine or xcms).
To the points:
JGK,
For targeted approaches, I do like this idea of redirecting the trash to waste. Also, there is apparently an add-on that I can get for my lc that will flip between columns while purging the previously used column... really cutting down on analysis time (are these common tools?). For high-throughput targeted approaches, this is really really getting interesting.
lmh,
(1) The response of the detector depends on ionising efficiency of the analyte. Since an analyte eluting late in your gradient isn't the same as one eluting early, it would probably have a different response irrespective of what gunk is being washed off by the solvent later in the gradient.
...
(3) Some of the changed signal and changed sensitivity later in the gradient is due to the solvents. Electrospray generally works better with lots of organic than pure water.
I have done an experiment with a few analytes in runs of differing %ages of isocratic organic; this really helps one understand ionization efficiency. Prior to doing these injections, I'd had trouble convincing another student that he couldn't say much about the relative abundance of analytes within a run. When one sees the abundance of the same analyte vary so widely, it brings the point home.
(4) The relationship of sensitivity to coeluting gunk depends on your instrument, and several mechanisms. Coeluting gunk can suppress ionisation in any instrument....
To me, this is one of the most interesting challenges for people working in ms. I suspect the answer is to just do the work, but if one is comparing the abundance of an analyte across samples and the composition of the matrix varies between samples, and a coeluting (unseen b/c of MRM approach) compound suppresses/enhances material in some but not other samples, then wow. Bad inference.
This makes me wonder about the influence of RT changes on standards. Hmmm... do you use different standards to differentiate extraction efficiency and RT changes? Is their literature on the effects of RT changes in ionization enhancement/suppression? Is this a bigger problem with MeOH? Do people ignore the issue?