Could any expert please advise? I appreciate your great help.
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Can a leak cause an RT shift in some of the compounds and not in others? Also, can a leak cause a drop in response? In tuning report, there is acceptably low percentage of 18 and 28 peaks. Can this exclude a leak?If you changes the trap you should have cooled down the GC and vented the mass spectrometer. If you didn't, you may have damages the column phase by the air introduced. If you did, check all fittings, the graphite vespel ferrules may have shrunk, causing a leak.
Thanks for your comment. If the ion source is dirty, the EM voltage and repeller voltage should be high?! But they are not high in my instrument. Since I see non-TMS compounds, is there a possibility that there is moisture somewhere in the GC column path? I checked the septum (actually replaced it) and made sure the injector nut was tightened. Then run the QC. But the low response problem and non-TMS compound were still present. Do I need to wait for the column to equilibrate so that any remaining moisture inside the column is gone? In addition, is there a need for the new column to run in for better sensitivity?If replacing the column fixed part of the problem, the old column was very likely damaged. While the column was dying it probably bled a lot of high MW contaminants into the MS, and the source now needs cleaning. That shoud fix the loss of detectability.
Peter
I just replaced the septum, gold seal and liner. I also made sure that the injector nut is tightened.Unless you have very dirty and wet carrier gas, which will destroy the column quickly in any case, it is not likely that the underivatised compounds that you see are TMS derivatives that have subsequently hydrolysed. Far more likely they were never derivatised in the first place, due to a problem with the sample preparation. The reason that you see the non-TMS compounds with the new column is that the old one was absorbing them completely.
I would be very surprised if the sensitivity got any better as the column ran in - columns rather tend to run out.
If the ion source is dirty high EM and repeller voltages can bring the sensitivy back again, but they would need to be adjusted accordingly - the MS does not tune itself based on the performance with QC samples.
What is your inlet maintenance schedule in terms of changing septa and liners ?
Peter