I'm glad this is a recent thread. I'd really appreciate any thoughts on the following situation / help rationalizing what I've observed.
I'm performing SEC, and thought I understood my total permeation volume which we've loosely called the salt peak.
While working through the cause of an early (~5min) negative peak I discovered its actually from the prior injection - I have an "over-lap-o-gram"
I confirmed this by extending my run time an additional 10 min (30 --> 40min), which resulted in that negative deflection showing up between 30 & 40min and it was absent in the next run.
Injecting a 16% soln of acetone (v/v in diluent), (10 µL) confirmed the total permeation volume was seen in under 30 min (UV 220nm). If I had a positive peak after my salt peak I would think retention was through a specific interaction (adsoption) in addition to size exclusion, but I'm at a loss to explain why a negative peak is seen with UV detection. I understand that a more transparent zone is in the detector, but can't figure why.
Diluent is pH 5 acetate, while MP is 20 mM MOPS, 200 mM NaCl, pH 7.5, 7% ACN in water.
One thought is that a component of the sample or its diluent (perhaps ACN?) is displacing a mobile phase component in equlibrium with the packing (adsoptive interaction). The displaced molecules then elute with the "salt peak". The more strongly retained molecules are isocratically eluted by a RP mechanism, which accounts for their retention beyond the total permeation volume. I'm guessing ACN, because the zone is detected as a decrease in UV absorbance.
Any thoughts / corrections / suggestions?