Changing column diameter
Posted: Mon Dec 08, 2008 4:43 pm
by syfir
I am thinking about moving from a .25mm column to a .18. Would I have to actually change anything other than the column info in the method? Will the EPC simply work out the pressure differential on its own? Also, any ideas on the increased likelihood of overloading the column?
Posted: Tue Dec 09, 2008 6:55 am
by Peter Apps
What the gas controls can and will do depends a lot on what kind of GC you have.
The "loadability" of a column depends on the cube of the diameter; the 0.18 mm column will overload with between a third and a half of what you can put onto the 0,25 mm.
Posted: Tue Dec 09, 2008 4:48 pm
by JI2002
My instinct response is that column loadability depends on the amount of stationary phase of the column. If you keep the phase ratio the same for a 0.25 mm i.d. column and a 0.18 mm i.d. column, then the overlaod limit of a 0.18 mm i.d. column is about 52% of 0.25 mm i.d. column.
If you don't keep the phase ratio the same, it is possible for the 0.25 mm i.d. column and 0.18 mm i.d. column to have same loadability by increasing the thickness of the stationary phase for the 0.18 mm i.d. column.
As always use Agilent method teanslator software to figure out the parameters for the new column. EPC doesn't do it for you.
More to Conisder
Posted: Mon Dec 15, 2008 10:51 pm
by gpronger
There are a few other aspects to consider.
Upside is less flow into you mass-spec.
Downside, if you're using a pulse-type injection, for an equivalent transfer from the inlet to the column, you'll need more head pressure during the pulse. Verify that you have sufficient line pressure for the pulse.
Greg