How about basic buffers in positive mode?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
How about adjusting the pH to below 8-9, and operate the MS in positive mode? Possible or just poor ionization effiency? I am working with neutral compounds with low ionization effiency, but want to get rid of interfering substances.

Or could I adjust pH of the mobile phase during a run? Which is best?

if I understand correctly, you have to be pH 8-9 to get the separation adn wsnt M+.

I guess you may be still able to get Na adduct.

If not, do postcolumn addition
Excel

You loose some ionization efficiency, but it still works well. We went to pH 10 for analytes like amitriptyline and were surprised that we still got a good signal. The signal was around 2 to 3 times less than at pH 3, but this is not a big deal if you can get around other problems, such as ion suppression from interferences.
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry