bad recovery!
Posted: Fri Dec 05, 2008 10:35 pm
I have the follwing situation:
1- Fentanyl citrate injection solution (containing NaCl and pH 4.7)
2- Standard: fentanyl citrate prepared in water
Result of the Assay by HPLC: 108%
when standard is prepared in NaCl solution at pH 4.7 instead of water the result of the assay by HPLC is 100%
The same phenomenon was observed for a perfusion solution of Fluconazole:
1- Fluconazole perfusion solution diluted in mobile phase 1:1
2- standard prepared in mobile phase
Pesult of the Assay by HPLC: 106%
a lab formulation was prepared and we obtained the same result 106%
means that if the lab formulation is used as standard the assy will be 100%
shall we prepare the standard by adding NaCl at the same concentration of the sample? or make lab formulation to control the unknown sample?
we changed the column
we changed the whole method (using another mobile phase etc..)
the result are the same
we get the methods from some scientific papers and the authors attest that the methods are validated...
i didn't give the detailed chromatographic conditions used because in both cases we tried different conditions.
has anyone an idea why this occurs? or has someone experienced such phenomenon?
thank you for your help!
1- Fentanyl citrate injection solution (containing NaCl and pH 4.7)
2- Standard: fentanyl citrate prepared in water
Result of the Assay by HPLC: 108%
when standard is prepared in NaCl solution at pH 4.7 instead of water the result of the assay by HPLC is 100%
The same phenomenon was observed for a perfusion solution of Fluconazole:
1- Fluconazole perfusion solution diluted in mobile phase 1:1
2- standard prepared in mobile phase
Pesult of the Assay by HPLC: 106%
a lab formulation was prepared and we obtained the same result 106%
means that if the lab formulation is used as standard the assy will be 100%
shall we prepare the standard by adding NaCl at the same concentration of the sample? or make lab formulation to control the unknown sample?
we changed the column
we changed the whole method (using another mobile phase etc..)
the result are the same
we get the methods from some scientific papers and the authors attest that the methods are validated...
i didn't give the detailed chromatographic conditions used because in both cases we tried different conditions.
has anyone an idea why this occurs? or has someone experienced such phenomenon?
thank you for your help!