Repeat injection issues
Posted: Wed Jun 05, 2019 4:21 pm
by Pear
Hi,
I'm noticing a strange behaviour from my HPLC when I inject the same sample repeatedly.
I have my system calibrated with know standards at:
0.1 ppm
0.2 ppm
0.3 ppm
0.4 ppm
0.5 ppm
The calibration is very accurate and gives good results.
However, when I run any sample, for example I did my 0.3ppm today, repeatedly, my results vary.
Real results from today:
0.3ppm calibration result = 0.3098ppm
But when I run this sample 3 times, I get the following:
1: 0.3020 ppm
2: 0.2910 ppm
3: 0.2830 ppm
If I continue, the sample gets lower and lower. Even if I do sample > blank > sample > blank, the results are the same.
Can anybody offer any input? Thankful for any help at all.
Re: Repeat injection issues
Posted: Wed Jun 05, 2019 5:45 pm
by Rndirk
Welcome to the forum.
Can you tell us what compound you are analyzing?
If you reinject a calibration standard after repeated sample injections, does it give a lower area as well?
Re: Repeat injection issues
Posted: Wed Jun 05, 2019 6:28 pm
by Pear
Welcome to the forum.
Can you tell us what compound you are analyzing?
If you reinject a calibration standard after repeated sample injections, does it give a lower area as well?
Hi Rndirk, thanks for the reply.
I'm analysing formaldehyde. The calibration standard is the re-injection 'sample' that gives a lower result each time.
Re: Repeat injection issues
Posted: Wed Jun 05, 2019 7:33 pm
by uzman
If you are using NASH reagent for post-column derivatisation , you have to protect your reagent from light , it is light sensitive.
Re: Repeat injection issues
Posted: Wed Jun 05, 2019 8:15 pm
by Pear
If you are using NASH reagent for post-column derivatisation , you have to protect your reagent from light , it is light sensitive.
Hi uzman,
We do pre-injection derivitisation. The reagent is kept in the dark at all times.
Re: Repeat injection issues
Posted: Wed Jun 05, 2019 8:53 pm
by uzman
Hi Pear ,
Is your injection solvent weaker than your mobile phase ; if you are using a RP column ?
If not , you may try to dissolve your sample in mobile phase and inject.
Re: Repeat injection issues
Posted: Wed Jun 05, 2019 10:15 pm
by Pear
Hi Pear ,
Is your injection solvent weaker than your mobile phase ; if you are using a RP column ?
If not , you may try to dissolve your sample in mobile phase and inject.
Hi uzman,
my mobile phase is 0.1% phosphoric acid + ACN: 0.1% phosphoric acid + water (45:55), the sample is added to 9:1 ACN:water then injected.
Re: Repeat injection issues
Posted: Thu Jun 06, 2019 7:03 am
by uzman
Hi Pear ,
Your injection solvent is stronger than your mobile phase , you may try mobile phase as injection solvent.
Re: Repeat injection issues
Posted: Fri Jun 07, 2019 10:29 am
by Pear
Hi Pear ,
Your injection solvent is stronger than your mobile phase , you may try mobile phase as injection solvent.
Hi uzman,
Thanks again for your help. We tried injections using mobile phase as the sample solvent, yet we are still seeing a decrease in result every injection.
Scratching my head over this one.
Re: Repeat injection issues
Posted: Fri Jun 07, 2019 11:14 am
by uzman
Some more suggestions :
First , try to backflush your column with pure acetonitrile , accumulated junk on the column might be the reason.
You may also switch to methanol as your mobile phase , instead of acetonitrile.
Formaldehyde ( as methylene glycol , in aqeous solutions ) is more stable in methanol.
Re: Repeat injection issues
Posted: Mon Jun 10, 2019 9:13 am
by lmh
just out of curiosity, if you take two vials, put the same standard in both, put them both in the autosampler, but inject from only one (repeatedly), if after a load of injections you then inject from the other, is it low or normal? I.e., is this a stability problem, or is it connected with making injections from the vial?