Jumping of retention time with comparable reponse
Posted: Fri Dec 05, 2008 12:51 am
Hello,
I am running 3 differnt samples preps with the same compoent of interest, but as run progresses the retention time would jump from the usual 12 mins to 25min the next injection but the response/area is comparable. This happens to all 3 preps and standards as well.
I suspect it could be my mobile phase changing in composition over time (it's a 45 hour run and the jumping started about 6 hours into the run) as my MPA is 100% water with minute TFA and MPB is 100% ACN with minute TFA, or it could be the pH of my mobil phases having an effect on the column, ie. hydrolysis of the bonded phase. How should I troublshoot this problem?
p.s the column and sample heating are both on.
I am running 3 differnt samples preps with the same compoent of interest, but as run progresses the retention time would jump from the usual 12 mins to 25min the next injection but the response/area is comparable. This happens to all 3 preps and standards as well.
I suspect it could be my mobile phase changing in composition over time (it's a 45 hour run and the jumping started about 6 hours into the run) as my MPA is 100% water with minute TFA and MPB is 100% ACN with minute TFA, or it could be the pH of my mobil phases having an effect on the column, ie. hydrolysis of the bonded phase. How should I troublshoot this problem?
p.s the column and sample heating are both on.