-
- Posts: 3
- Joined: Mon Dec 01, 2008 11:35 am
Column- Phenomenex Luna C18-(2) 5 μm x 250 mm x 4.6 mm
Also used Hicrome and Waters columns, old and new columns, guard columns used and not used
Column temperature 25 oC
Mobile Phase Acetonitrile: deionised water (50:50).
Different suppliers and grades of ACN and water used
Flow rate 1.0 ml/min
Runtime 10 min
Injection volume 100 l
Detection UV at 210 nm
Samples are prepared in acid at pH 1.2. In all acid blanks and samples (dissolution samples including tablet) the ghost peak appears.
Acid makeup has been investigated without using pH meter and only glassware used. Different grades and suppliers of HCL and water have been used for this prep.
Different injector vials used
Different HPLC systems used
Different analysts
Solvent inlet filters on HPLC changed
No contamination peak from water blank, no contaminant peak for a zero mico L injection on HPLC
Systems cleaned down with ACN, Water, IPA
Any help would be greatly appreciated
