Hello!

I am new on this forum and would really like to hear some experiences with PL ELS 1000 detector, which we have in our lab. I am the first one who uses this detector in our lab. The sensitivitiy of the detector is just fine, but I got another problem. And that is the: evaporation of organic solvents (such as THF, CHCl3, ACN...). We have bottle traps for condensat, which are in the lab. The thing is that in spite of that, our lab is full of solvent vapour and many times i have problem with headaches.

What to do?What are u experiences?

Regards,
Tina

ELSD's (from any vendor) need to be vented! The trap bottle will never collect all of the vapors.

The instructions from your ELSD manual should instruct you to have the exit line from the solvent trap venting to a fume hood or extract vent.

I'm surprised that you have only had headaches so far.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry