How to connect the EPC to QTOF Ultima?
Posted: Fri May 31, 2019 7:02 pm
by iyanachk
Can anybody advise me how to connect the cables of Micromass QTOF Ultima API to the embedded computer (EPC) board. The board is TTP-20221-ATOM. There are two co-axial cables, one from preamplifier of the MCP detector and another from "PUSHER SYNC". On the board there are three co-axial sockets. The middle one is plugged, but the other two are not labeled and are the same, so I am not sure which cable goes where.
Thank you very much!
Re: How to connect the EPC to QTOF Ultima?
Posted: Tue Jun 04, 2019 11:31 pm
by mhr311
Coax from preamp goes to TDC input. Coax from pusher goes to TDC trigger. The trigger connector is usually the outer bnc plug.
Posted: Fri Jul 05, 2019 1:18 pm
by iyanachk
Thanks a lot, mhr311. Below is a link to a picture of the back of the EPC. The bnc sockets are labeled with 1 and 2. Can you point which one might be the trigger connector?[img]"https://drive.google.com/file/d/18Jg7R_bhVJYSoOiwpq1yh4Rt2YoLN7br/view?usp=sharing"[/img]
Thanks again!