We are a Non-regulated Lab. I have the authorization to change methods. Our HPLC:
Waters components: In-line Degasser, 1515 Pump, 717+ Auto sampler, 2414 RI Detector
Column IC-Pak Ion Exchange 7.8 x 150 with a Shodex SH-G Guard Column
Mobile phase: 0.5 mM H2SO4
We are currently running a 10 minute method; I have access to an equivalent 20 minute method as a last resort.
Our components of interest include (in order of elution):
Dextrin, Maltotrios, Maltose, and Glucose
Lactic Acid, Glycerol, and Acetic Acid
Ethanol
The Standard & Check Standard are acceptable.
A few samples have bad Chromatograms:
- Peaks shifting together early in retention (The maltotrios moves into Dex or Maltose)
- Peak Tailing or broadening early in retention (Maltotrios can be completely lost)
- Mystery peaks late in retention (In the long gap between Acetic Acid and Ethanol)
Here is my logic:
Since the Standards look good the issues must be sample related
It’s possible some samples are a stronger solvent than the mobile
Either known peaks or unknown peaks could be the problem
Decreasing Injection Volume?
Some samples have very low concentrations of the analyte of interest. (The standard has approximately 0.3%) Resolution priority should go to the acids, even if resolution suffers in the sugars. I could decrease the injection volume from 10 uL to 5 uL, but, I don’t want to lose the acid peaks into the baseline. Plus I still wouldn’t know which peak was the problem…
Spike the Standard?
So I would also like to "spike" different vials of my standard with some individual components to see if high concentrations of one component are interfering with other peaks. This would also identify if the shifting maltotrios really is maltotrios. It may be possible I have a totally different sugar in the area; I tried a sample from a different feed stock, sucrose co-eluted in my other sugars.
Questions:
1. Does my logic make sense?
2. What concentration of components would you start with when spiking a standard?
3. Is it acceptable to inject 10 uL first, and only re-inject problem samples at 5 uL?
4. Can you refer me to a resource for carbohydrate separation?
Jade Barker