CTC PAL - crushing syringes on power off?

Discussions about GC and other "gas phase" separation techniques.

6 posts Page 1 of 1
I have acquired used GC and LC PAL autosamplers. Both of them when power is turned off let the injector fall down. If syringe was installed it would be broken in no time.

Is this normal? Or something (springs? tension cord? brake?) needs to be adjusted/replaced in the injectors?
Hi

When PAL sampler is off injector should slowly fall down, depends how good tension cord is.
As there are 2 tension cords I suspect you've got at least one completely worn out.

Regards

Tomasz Kubowicz
When I power off the PAL, I like to remove the tray, then place my finger just behind where the syringe would extend through the syringe guide, turn unit off, then gently move the whole arm and guide it to the hole in the tray holder and "park" it there. Syringe safe.
Regards,

Christian
Thanks for advice. I will probably need to replace the tension cords then.
We have all of ours serviced each year. Cords replaced, rods and gears cleaned, etc. Think we've got 10 of them.
Regards,

Christian
As recommended, the tension cords need to be replaced.
6 posts Page 1 of 1

Who is online

In total there are 24 users online :: 1 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry