CTC PAL - crushing syringes on power off?

Posted: Mon May 20, 2019 8:00 am
by fzawadiak
I have acquired used GC and LC PAL autosamplers. Both of them when power is turned off let the injector fall down. If syringe was installed it would be broken in no time.

Is this normal? Or something (springs? tension cord? brake?) needs to be adjusted/replaced in the injectors?

Re: CTC PAL - crushing syringes on power off?

Posted: Mon May 20, 2019 2:21 pm
by tkubowicz
Hi

When PAL sampler is off injector should slowly fall down, depends how good tension cord is.
As there are 2 tension cords I suspect you've got at least one completely worn out.

Regards

Tomasz Kubowicz

Re: CTC PAL - crushing syringes on power off?

Posted: Mon May 20, 2019 8:09 pm
by cjm
When I power off the PAL, I like to remove the tray, then place my finger just behind where the syringe would extend through the syringe guide, turn unit off, then gently move the whole arm and guide it to the hole in the tray holder and "park" it there. Syringe safe.

Re: CTC PAL - crushing syringes on power off?

Posted: Fri May 24, 2019 12:45 pm
by fzawadiak
Thanks for advice. I will probably need to replace the tension cords then.

Re: CTC PAL - crushing syringes on power off?

Posted: Fri May 24, 2019 6:56 pm
by cjm
We have all of ours serviced each year. Cords replaced, rods and gears cleaned, etc. Think we've got 10 of them.

Re: CTC PAL - crushing syringes on power off?

Posted: Fri May 24, 2019 8:41 pm
by Jake
As recommended, the tension cords need to be replaced.