I am developing an External Standard test method for the determination of residual solvents in an API.
The sample preparation procedure is as follows:
Transfer about 25 mg of the sample, accurately weighed, to a 20-ml headspace vial, add 5.0 ml of water, apply the stopper, mix.
Vigorous shaking of the headspace vial with hand can not get the sample dissolved at ambient temperature. But the sample is dissolved, when the vial is incubated at 80 ℃ for 30 min in Agilent G1888 HS sampler.
I remember the USP <467> requires that the article under test be dissolved in diluent. I have no idea whether the proposed sample preparation above meet the USP requirement. I hope someone can help clarify this.
Many thanks in advance.
Terry