Waters 2695 / Agilent 1100 max column temperature

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi everyone,

We are looking to buy an HPLC system to run some carbohydrate analyses, and the protocol calls for a column temperature of 85°C. I have seen several papers saying they have used this protocol and temperature using a Waters 2695 or Agilent 1100 system, however the manufacturers' specifications indicate temperatures up to 60°C and 80°C, respectively.

Is there any known way to get the column compartments to heat up higher than the manufacturer's specification?

Thank you for your help!
Corentin @ C16 Bio
Not to my knowledge, but it is highly likely they used an external column compartment (third party) with these systems.
Hi
Waters has an external column heater. Max temp is 150 C
We just used a stirring beaker full of filtered water at 85 Celsius as the mobile phase supply for 'Simple Sugar analysis' (~40 years ago).
4 posts Page 1 of 1

Who is online

In total there are 391 users online :: 0 registered, 0 hidden and 391 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 391 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry