Hi All,
Can anyone tell me why - when I have the split flow as OFF there is still flow out through the split line (equivalent to the flow at the split ON event?
Thanks in advance
WK
I'm Sorry I Haven't A Clue - Just A Minute - The Unbelievable Truth

It's because during splitless operation the bulk of the gas flow bypasses the inlet, rather than being shut off as it was with the older inlet pneumatics. There is a diagram in the inlet manual.

This arrangement is a consequence of back-pressure control of the pressure in the inlet, which requires mass flow control of the flow to the inlet. When this arrangement was first commercialised the advantage that it can maintain inlet pressure even when there is a leak was punted as a major step forward, but this is true only for high flow rates, and with wide connecting lines, and if you have those an "old fashioned" front-pressure control system is also robust to leaks.

As far as I know all GCs now have back-pressure control.

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry