They definitely are. Try washing with ammonium acetate or ammonium formate in 100% water; hopefully you have an amino phase that can handle that.Is there a possibility that TFA is accumulated on the surface of the amino column and extensive wash is needed to resume chromatographic performance???
Thank you very much for your reply...I hope that disorption of ligand phase and sample gunk are not the cases here because I haven't ingected any "real samples" into the column. In addition I am working only with standards for approximately two months and at pH range 3-6. The truth is that I have tryed tetrahydrofuran to wash out any tightly bounded anions. A procedure employed for graphitized carbon columns..as far as 100% H2O, I am not sure if my column will handle such conditions, so I will communicate with the manufacturer.I suppose reasons for the reduced performance:
- sample gunk accumulated on column
- desorption of ligand phase
- anions are tightly bound to ligand
I think different column manufacturers will have different cleaning
procedures for their NH2 phases. So you may want to ask the
manufacturer.
For our NH2 phase, we would recommend 50mM NH4AcOH or (formate).
The less organic - the better (but not every NH2 phase can handle
100% aqueous eluent - so be mindful of that).