lmh » Fri Nov 07, 2008 1:25 pm
(1) About picking up noise in the baseline as a signal:
Don's right, both about baselines and the human eye. For human-eye reasons, I don't trust manual integration (and haven't done it for years). But personally I also set my integrator less sensitive, so it won't find baseline noise. If you have it more sensitive, there will be peaks all along the baseline. This has the disadvantage that if your retention times wander (and I won't tell you why I'd worry about a thing like that... never happens, does it?), there is a myriad of spurious peaks just waiting to be misidentified...
(2) Do we include zero?
JGK correctly and usefully quotes the FDA guidelines. Now, these guidelines suggest we should use the simplest regression that we can for the calibration curve. If you force through zero, you are using the equation y = mx. If you don't force through zero, you are using y=mx+c. Therefore the guidelines actually hint we should force through zero, though I'm sure that wasn't in the minds of those who wrote them, nor is it necessarily routine practice.
But does it matter? The guidelines also state that the calibration points must be within +/-15%, and that the LLOQ point must be within +/- 20%. This means that however you arrived at your cali curve, any measured points should be within +/-15% unless they're near the LLOQ, in which case they may be +/-20%. Frankly, whether you forced through zero or not, you must arrive at a line that fits the points, or you won't fit the criteria! If forcing through zero makes a significant difference to the accuracy of the LLOQ standards, then there is a serious problem with the calibration curve. It's more important to solve the problem than to debate whether to include zero.
(3) Why is the day-to-day precision lower if you force through zero
If you measure something in the middle of a calibration curve, the precision of this measurement depends on errors in the measurement, and errors in the calibration curve. The error in the middle of a 2-point calibration curve is the average of the errors at either end. If you force through zero, then the error at one end becomes zero, so the precision in the middle is improved. But the fact remains, unless you have all the calibration points, you cannot be sure that the accuracy is good over the whole range. The standards at the LLOQ also provide a genuine measurement of the precision expected at the LLOQ, and important thing.
(4) FDA guidelines and ignoring points
The deeply scary bit of the FDA guidelines is the fact you are allowed to throw away (more or less at will) a third of calibration and QC points. This has two consequences. Firstly it means the +/-15% limit isn't really 15%. It can actually be substantially higher, because you can trim outliers and still pass. Secondly it has consequences for how many replicates you must do of each sample. If you expect 33% of QC samples to be outside an acceptable window, then 33% of real samples will also be outside the window. Unfortunately, for real samples, in contrast to QC samples, you don't know where the window should be. Scaled up to 100 real samples, how many replicates of each do you need to make sure that none of the 100 samples is outside a window of +/-25%?
In my view, this license to delete points is deeply mischievous, and undermines the guidelines completely. I sincerely hope no one is seriously doing it.