i have a puzzling baseline shift occuring randomly on two of our HPLCs. i am finding it hard to find a pattern or pin it down to one cause.
the shift is slight (1-10mau) but is enough to interfer with our impurity analysis and %recoveries.
I did a forum search and found an entry (fri,mar 04 2005 by jzt) with a very similar effect although not as large a shift. But we are getting the same "cliff" effect which when it arises at the rt of our peaks is very troublesome.
In our case the shift has occured in two different instruments with the same set up(1200, DAD,ALS, Quart pump), which were installed at the same time. in both cases the shift began 4-5 months after 1st use and up to present.
It occures with completly different methods, different columns, different wavelenghts, different chemists. the hplcs are even in different rooms!
but it occurs periodically from time to time, often not for weeks.
However they are all isocratic methods. the only common factor not including electricity supply!
We aquired two Agilent 1200 systems a year ago, oct 07.
We had an agilent engineer out to take a look, he changed seals,needle seat and needle but it made no difference.
Has anyone any ideas? or has anyone had similar issues?
many thanks,
derek