Page 1 of 1

Column deterioration by metal chelate samples?

Posted: Wed May 01, 2019 8:53 pm
by Chromma
Hi all,

I'm currently running purity and assay tests on an organic metal chelate (complexed with iron, 5 wt%). I'm using a C18 endcapped stationary phase and ACN/water mobile phase with TFA (gradient 5% --> 95% organic) combined with UV detection. However, I'm experiencing increase of area response and decrease of resolution over time. I was wondering whether this might perhaps be caused by stationary phase saturation/deterioration by iron over time?

Does anyone have experience with this problem or with analysis of organic metal chelates by RP HPLC in general?

Thank you for your response.

Re: Column deterioration by metal chelate samples?

Posted: Thu May 02, 2019 8:12 am
by Gerhard Kratz
No matter how is your endcapping - double, 100% or best in the world, still there are free -SiOH groups on the surface and with your metal ions they are happy to bind and that will lead to the fact that you need to purchase a new column.
I would use another mode than RP18.
Good luck

Re: Column deterioration by metal chelate samples?

Posted: Sat May 04, 2019 1:13 pm
by Chromma
Thank you for your response! Which alternative chromatographic mode(s) to RP C18 would you suggest?

Re: Column deterioration by metal chelate samples?

Posted: Mon May 06, 2019 5:30 am
by Gerhard Kratz
IC could help. https://www.shodex.com/en/dc/07/04/06.html here maybe you will find some answers.