Page 1 of 1
Flame Ionization Detector
Posted: Tue Nov 18, 2008 9:17 am
by Andronof
Good morning to all
I am searching for tables about flame ionization detector response factors referred to methane or propane.
Could someone help me?
Thx!
Posted: Tue Nov 18, 2008 2:52 pm
by GasMan
Andronof
Check out page 8 of the ASTM method D5443.
Gasman
Posted: Wed Nov 26, 2008 12:51 pm
by dokwill
and astm d 6729
responce factor
Posted: Wed Nov 26, 2008 1:40 pm
by praveenpaliwal
and astm d 6729
Need not to find any book. Weigh accurately Methanol and propanol and note down the weight. run on GC . Find out the area %
Now % wt / % area = Responce factor.
Posted: Thu Nov 27, 2008 12:09 pm
by dokwill
i will be tomorrow in lab
i will check in my gc.
wait here
Re: Flame Ionization Detector
Posted: Thu Nov 27, 2008 5:24 pm
by Consumer Products Guy
Good morning to all
I am searching for tables about flame ionization detector response factors referred to methane or propane.
Could someone help me?
Thx!
Need not to find any book. Weigh accurately Methanol and propanol and note down the weight. run on GC . Find out the area % Now % wt / % area = Responce factor.
Methanol and propanol are different than methane and propane.
Posted: Fri Nov 28, 2008 12:21 pm
by dokwill
rf to varian fid
in astm d 6729, 5441, 5134
metane 1.2000
propane 1.0500
Posted: Fri Nov 28, 2008 3:51 pm
by Peter Apps
So much for the FID being a non-selective detector !
Peter
Posted: Sat Nov 29, 2008 9:17 am
by dokwill
this we use every day