Reproducibility Problem
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
I have problems of reproducibility, in an Alliance 2695, have mounted 3 systems of 6 injections, and always in the three systems the first injection is the one that does not play like other 5. im not have idea what it is.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
It's hard to give a specific answer without a lot more information about your method, but just as a guess: If your separation involves a gradient, the pre-equilibration time can have an impact on results. In an automated system, the equilibration time between injections is constant *except* for the first injection of the day. My practice has been to always make the first injection a blank or "dummy" to avoid that problem.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
We ALWAYS did that; likely learned that from you in the late 1970s !!!My practice has been to always make the first injection a blank or "dummy" to avoid that problem.
We wrote that into our test procedures so there would be no confusion, and for our cGMP compliance.
-
- Posts: 154
- Joined: Thu Jan 20, 2005 11:27 pm
Tom Jupille and Consumer Products Guy:
Great Advice! I must have learned this from reading your advice over the years.
I generally include 1 - 4 equilibration injections of standard prior to system suitability to condition the system, depending upon the assay.
Also, I suggest adding time to the re-equilibration at the end of the chromatogram or between injections, if possible.
Great Advice! I must have learned this from reading your advice over the years.
I generally include 1 - 4 equilibration injections of standard prior to system suitability to condition the system, depending upon the assay.
Also, I suggest adding time to the re-equilibration at the end of the chromatogram or between injections, if possible.
4 posts
Page 1 of 1
Who is online
In total there are 569 users online :: 0 registered, 0 hidden and 569 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 569 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 569 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science