Analysis of benzoic and salicylic acids
Posted: Sun Apr 14, 2019 3:43 pm
by Andre Orci
Hi everyone!
Can i perform the analysis of pure benzoic and salicylic acids in an Agilent HP-FFAP column (FID detector)?
If yes, what is the methodology for the analysis (sample preparation, GC parameters...)?
I try to dilute the acids in acetone (3% w/w) but the analysis had a poor resolution.
(I also have an HP-INNOWax column. If this column provides better results than the FFAP, i would like to know the methodology for this one).
Re: Analysis of benzoic and salicylic acids
Posted: Sun Apr 14, 2019 4:31 pm
by HPLC chemist
Yes. But HPLC is even better and has replaced thus GC analysis. See attached for GC chromatographic conditions;
https://www.agilent.com/cs/library/appl ... 4582EN.pdf
Posted: Sun Apr 14, 2019 5:20 pm
by Consumer Products Guy
Yes, HPLC using reverse phase and low pH mobile phase and UV detector is typically easier.
But there were instances when we assayed these by GC, but we derivatized with BSTFA and used a non-polar capillary.
Posted: Mon Apr 15, 2019 12:30 am
by rb6banjo
I've done this sort of thing by solid-phase extracting them out of the aqueous matrix and derivatizing them with MeOH:BF3. After quenching the reagent with aqueous sodium phosphate, analyze for the methyl esters by headspace SPME.
Posted: Mon Apr 15, 2019 6:02 am
by Henrylozano
Hello, I think that you can perform one method from direct injection without Derivate any components. Can you send me the list of compounds and I can make one method that you can to essay?
Posted: Wed Apr 17, 2019 7:28 pm
by James_Ball
Would derivatizing a 3% solution of benzoic acid require quite a bit of derivatizing agent?
Posted: Wed Apr 17, 2019 9:00 pm
Can i perform the analysis of pure benzoic and salicylic acids in an Agilent (GC using) FID detector?
I think his samples are pure substances. I'd dilute a small amount of sample in DMF or pyridine (my first tries) and then take 250ul and add 125ul BSTFA and mix, then inject.
Even if I had 3% solution in water, I would dilute 0.25 grams sample in DMF or pyridine and then take 250ul and add 200ul BSTFA and mix, then inject. We would use that same sample dilution on samples containing 90% water to assay for analytes 0.1% or lower in the sample.
Posted: Thu May 16, 2019 12:17 pm
by MSCHemist
For esterfication of larger acids I like mild methanolysis. https://www.ncbi.nlm.nih.gov/pmc/articles/PMC2817593/
I've used HCl methanol and done either the overnight 40 deg C or 30 minutes at 90 deg C. It works on diacids like succinic, fumaric etc as well it does hydroxy acids but some give me activity issues with the hydroxyl (citric/tartaric).
Posted: Thu May 16, 2019 3:05 pm
Funny, even though we made methyl esters of fatty acids thousands of times, we never bothered for benzoic or salicylic acids.
Like I said before, we ether used BSTFA or simply used HPLC.