My lab is currently trying to get a water method that has been used on a 5890 to work on a 7890 GC, but are having some problems with tailing on the first peak. We use a double tapered gooseneck liner on the 5890, but have tried a couple of different liners on the 7890. The only reason we have not kept to using the double taper liner on the 7890 was because our RSD (needs to be at 5% or lower) was failing. Our RSD was in the upper teens. If we used a 4mm liner with wool, our RSD passed but the tailing on the first peak was around 1.7-2.2, when we need it to be below 1.5. I have also used the following liners, low pressure dropliner with wool and gooseneck splitless liner, to no success.
Here is some information of how the 5890 is setup:
Column: Wax 15m x 0.53um from J&W
Double taper liner
Injection volume: 1uL
Injector: 210C
Detector: 250C
We have tried a lot of different things on the 7890 to fix the tailing problem, but nothing seems to help. We have tried a 30m column with a higher pressure, we have tried injecting just 0.5uL, we have increased the split ratio, we have decreased our initial oven temp.
I am hoping someone can help guide me in some sort of direction.
Thank you very much