I have a Agilent 3000 micro GC with OV-1 and Molecular Sieve 5A columns. Is it possible to detect and separate N2O and C3H6 on OV-1 column?
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Hineelambj,
The Nist site shows a retention index for that at 186 on an SPB-1 which means between methane and ethane (just before ethane.) If I am not mistaken, OV-1 would be very similar retention.
Please advise if this is the case. Also, what levels are you looking for?
Best regards.
Hi...Hineelambj,
The Nist site shows a retention index for that at 186 on an SPB-1 which means between methane and ethane (just before ethane.) If I am not mistaken, OV-1 would be very similar retention.
Please advise if this is the case. Also, what levels are you looking for?
Best regards.
Thanks a lot for the reply. Methane will get detected on MS-5A. However this is not my concern at the moment. Iwant to detect propylene (2000 ppm and less, maybe upto 100 ppm) and N2O (100 ppm and above). Is it possible to detect these two gases on this particular column, and if yes is there separation possible.
Regards
Neelam