Small peak problem
Posted: Wed Nov 05, 2008 11:25 pm
Hello friends-
I am running a two-component matrix. Two analytes. One is Octinoxate (OMC) and the other is Avobenzone (AVB).
Octinoxate comes out in 4.8 minutes and the Avobenzone comes in 6.5 minutes.
Anyway, the first couple of sets of analysis that I have run, my OMC peak had much higher response than the AVB peak. (In assay analysis OMC should be around 6%, and AVB is about 2%).
The last two times it's the opposite!- The AVB peak (The second peak that comes in 6.5 minutes with 2% analyte) is much higher. It is very strange!
The peak shapes are good. No problem whatsoever with the peak shapes-
The area counts do not make a difference and the results are acceptable. They are in range.
My detector intensity is ok, and my lamp intensity is good as well. My system suitability has passed in every single time. - I've checked it.
The lamp has only 130 hours on it.
The detector range changes from 310 nm to 357 nm in 5.5 minutes, so it can detect both analytes at given retention times,
My HPLC is Dionex P680 with ASI 100 and PDA 100 Photodiode Array detector
My column is RP Supelco 15 cm x 4.6mm x 5 mm
My flow rate is about 1 ml/min
My mobile phase composition is 90: 8: 2 (Methanol: Water: 0.5% H3 PO4)
My Chromatography software is Chromeleon
Solvents look good- I filter them on regular basis- But it's been a mystery problem- Thought anyone would have an idea what's happening?
Now that OMC peak is smaller- The software doesn't integrate the first peak properly and my calibration for the first analyte is not perfect. I changed the signal to a smaller value, so it can integrate it properly.
I can run 3-point calibration and the data looks ok for the first analyte. It is perfect for the second one.
However; If I ran 4 point calibration, the highest calibration point for the first analyte comes above the linear regression curve, and henceforth; my results on the first analyte comes lower than it should be. No problem with the second analyte calibration.
I ran recoveries before and they came out ok- Not between 98-102% but consistent-
I have no idea why this is happening-
Any word of wisdom why this is acting weird?
I would appreciate it very much
Thank you
I am running a two-component matrix. Two analytes. One is Octinoxate (OMC) and the other is Avobenzone (AVB).
Octinoxate comes out in 4.8 minutes and the Avobenzone comes in 6.5 minutes.
Anyway, the first couple of sets of analysis that I have run, my OMC peak had much higher response than the AVB peak. (In assay analysis OMC should be around 6%, and AVB is about 2%).
The last two times it's the opposite!- The AVB peak (The second peak that comes in 6.5 minutes with 2% analyte) is much higher. It is very strange!
The peak shapes are good. No problem whatsoever with the peak shapes-
The area counts do not make a difference and the results are acceptable. They are in range.
My detector intensity is ok, and my lamp intensity is good as well. My system suitability has passed in every single time. - I've checked it.
The lamp has only 130 hours on it.
The detector range changes from 310 nm to 357 nm in 5.5 minutes, so it can detect both analytes at given retention times,
My HPLC is Dionex P680 with ASI 100 and PDA 100 Photodiode Array detector
My column is RP Supelco 15 cm x 4.6mm x 5 mm
My flow rate is about 1 ml/min
My mobile phase composition is 90: 8: 2 (Methanol: Water: 0.5% H3 PO4)
My Chromatography software is Chromeleon
Solvents look good- I filter them on regular basis- But it's been a mystery problem- Thought anyone would have an idea what's happening?
Now that OMC peak is smaller- The software doesn't integrate the first peak properly and my calibration for the first analyte is not perfect. I changed the signal to a smaller value, so it can integrate it properly.
I can run 3-point calibration and the data looks ok for the first analyte. It is perfect for the second one.
However; If I ran 4 point calibration, the highest calibration point for the first analyte comes above the linear regression curve, and henceforth; my results on the first analyte comes lower than it should be. No problem with the second analyte calibration.
I ran recoveries before and they came out ok- Not between 98-102% but consistent-
I have no idea why this is happening-
Any word of wisdom why this is acting weird?
I would appreciate it very much
Thank you