empower 1 and UPLC
Posted: Wed Nov 05, 2008 8:30 pm
by chromochat
Is UPLC compatible with empower1?
Posted: Wed Nov 05, 2008 9:03 pm
by DR
Yes - there will be some ICOPs to install, but yes, it is.
(thread moved to "Data Systems")
Re: empower 1 and UPLC
Posted: Wed Nov 05, 2008 9:59 pm
by chromochat
will you explain what is ICOPs?
Posted: Thu Nov 06, 2008 3:26 pm
by DR
ICOPs are instrument control programs.
Acquity systems were not around when Empower was introduced, so for any Waters instrument that is either uncommon, or newer than Empower, there is a set of programs that gets run on the server and on the LAC/E attached to said instruments that allows the instrument to function.
If you'r in a client/server environment involving Citrix servers, the ICOPs have to go there too. Your Empower admin. should be up on all of this.
If you're more or less alone, running a local installation ("fat client"), you will need the ICOPs on your LAC/E as well as on your computer.
The ICOPs come with the isntrument, can be downloaded from Waters etc. etc.
Posted: Thu Nov 06, 2008 5:15 pm
by chromochat
thanks DR!